Would something like THIS work for what you have in mind? If nothing else, the same error would be in each of your readings.
I use these for rough temp readings, such as the underside of the roof in the observatory, drive motor temp, heat issues with Observatory Computer ...
Dale Eason
Reged: 11/24/09
Loc: Roseville,Mn.
Re: Instrumenting a Dob with Temperature Probes new [Re: Tom and Beth]
      #5943936 - 06/27/13 10:37 PM
The problem to watch out for with those non-contact thermometers is that glass even uncoated is reflective in IR. So they don't read the temp of the glass but what is reflected by it. I solved that problem by putting a piece of tape on the back of the mirror and pointed it at that.
Dale Eason
careysub
Reged: 02/18/11
Loc: Rancho Cucamonga, CA
      #5944143 - 06/28/13 01:55 AM
IR thermometers are useful, I have one that I have used to monitor open back Dob mirror temps.
They do not however take air temperatures, an important aspect of monitoring cooling design; air temps also allows monitoring the air circulation - it is not only directly relevant measurement, it is also a proxy for monitoring the air flow through the system.
Also, not all parts of telescope systems are directly accessible to IR monitoring (closed Dob boxes, telescope tubes).
careysub
Reged: 02/18/11
Re: Instrumenting a Dob with Temperature Probes new [Re: don clement]
      #5944149 - 06/28/13 01:59 AM
I am sure there are five figure lab units that also make good voltmeters.
Ah the late lamented HeathKit! The website assets "the news of my death is greatly exaggerated" but that is so far the only sign of life.
FHarry
Reged: 04/20/07
Loc: Bavaria
      #5944170 - 06/28/13 02:38 AM
why not use the one-wire sensors? Super-small in size and cheap as well. If you would not like to put the necessary sensors to a bus-cable (actually, a simple cable like the Cat5 or so and all sensors connected as a bus to this cable) then you even could use the one-wire buttons called iButton. Sensor with battery in a small housing, like a coin. So you could equipp your scope with this "buttons" wherever you like. In the end, you pick them out and put them into a little reader unit and read the data. Buttons can be configured to take a measurement all n-minutes or so.
don clement
Vendor (Clement Focuser)
Reged: 02/02/11
Loc: Running Springs, California
      #5944486 - 06/28/13 09:52 AM
Yes Kiethley instruments come to mind http://www.keithley.com/
I am sure there are many others that believe that the vacuum tube is dead. There are also the few of us still channeling those dead vacuum tubes for the clairvoyant sounds of long deceased vinyl.
Don Clement
Pinbout
Reged: 02/22/10
Loc: nj
      #5944502 - 06/28/13 10:03 AM
Don Clement
but the day has long been dead that you drive to the 7-11 to check out if the tubes are blown.
I still have my gt101 accoustic amp
careysub
Reged: 02/18/11
Re: Instrumenting a Dob with Temperature Probes new [Re: FHarry]
      #5944556 - 06/28/13 10:39 AM
Interesting. I did not know about the one-wire technology.
This sort of low cost, low performance digital microlan looks promising. Here is a good introductory page from a hobby component provider:
https://www.hobby-boards.com/store/pages.php?pageid=4
I found this site interesting:
http://www.digitemp.com/
Temperature monitoring software for Linux, I have two EEPC netbooks running EEBuntu that might be just the thing for this. This would solve the controller/monitor problem with looked like the most costly/time consuming part of any digital solution.
The iButtons look a lot more expensive on a per sensor basis.
Are there any books out there providing a good grounding in 1-wire network design and set-up; e.g. how to use these little digital switches and other network components I see? Something along the lines of the Arduino Cookbook for example?
This looks like a cheap and convenient and versatile enough technology, with a low enough cost of entry, to merit closer examination.
freestar8n
Reged: 10/12/07
Re: Instrumenting a Dob with Temperature Probes [Re: careysub]
      #5944591 - 06/28/13 11:00 AM
The ds1820's that I used in my sct cooling study mentioned earlier also operate in one-wire mode, allowing multiple sensors to be connected on the same input and individually addressed. The newer version, ds18b20, has 12-bit precision, or 0.06C. This sensor family worked well for the specific application of measuring temperature at different points in a telescope, with no analog signals floating around. But I was comfortable writing custom PIC code for the whole thing.
Frank
Pinbout
Reged: 02/22/10
Loc: nj
Re: Instrumenting a Dob with Temperature Probes new [Re: freestar8n]
      #5944722 - 06/28/13 12:35 PM
shouldn't be a problem to wire into an arduino or even a basic stamp?
Reged: 06/16/04
      #5945086 - 06/28/13 03:52 PM
My thinking
Repeatability between probes is very important. So... When it comes to repeatibility as well as accuracy, it it hard to beat the old fashioned thermocouple. Thermocouple wire is relatively cheap so you can make your own probes. The voltage output is quite low so an amplifier of some sort is needed. I have an old Fluke multimeter with a seperate module that fits any multimeter and outputs 1 mv per degree either F of C. The Fluke has a millivolt scale that reads to 0.1 mv so this is 0.1 degree.
With other sorts of sensors, the repeatability is questionable between sensors.
As far as using an IR device to measure the temperature, the amount of IR emitted by a body depends on both its temperature and its emissivity so one cannot make accurate measurements of different surfaces without calibration.
don clement
Reged: 02/02/11
Re: Instrumenting a Dob with Temperature Probes new [Re: Jon Isaacs]
      #5945264 - 06/28/13 05:41 PM
Thermocouples are non-linear. One must provide a lookup table to linearize thermocouples. Some of the old Fluke DVM did do that for K thermocouples IIR. There is a simple way to match the output on LM35 linear temperature sensors using 10K trim pots.
See: http://www.investigacion.frc.utn.edu.ar/sensores/temperatura/AN-460.pdf
Don Clement
freestar8n
Reged: 10/12/07
Re: Instrumenting a Dob with Temperature Probes new [Re: Pinbout]
      #5945266 - 06/28/13 05:41 PM
Yes - many chips and environments would work for development - but my main point is that custom work is helped by being able to write code specific to the needs of the project.
As for accuracy, when it comes to air measurement I think a key limitation is a good shroud for the sensor so it isn't heavily affected by radiation from nearby surfaces and the sky - and is a true measure of the local air temperature rather than the radiative properties of the sensor in its environment.
Frank
Reged: 06/16/04
      #5945357 - 06/28/13 06:44 PM
Don Clement
It is true that thermocouples are nonlinear over their full range but the full range of a K type is up to 1370C. Ktypes are reasonably linear to a 1000C. In the range of interest here they are linear and in any event these measurements are relative rather than absolute.
don clement
Reged: 02/02/11
      #5945444 - 06/28/13 07:31 PM
Omega is a good place to get thermocouples,connectors and other types of temperature probes. http://www.omega.com/temperature/tsc.html
Don Clement
Reged: 06/16/04
      #5945459 - 06/28/13 07:40 PM
Don Clement
That's where I get our stuff for the lab. We have a wonderful old digital thermocouple reader, it's 12 channels, switchable, reads to 0.1 degrees. But I think a data logger would be slick for this app. There are some inexpensive USB data loggers out there that would do the job.
careysub
Reged: 02/18/11
      #5945943 - 06/29/13 03:03 AM
Don Clement
Putting the Omega type K specs into a spreadsheet I see that average mV/C over the range of -10 to 50 C is 0.0403; over the more restricted range of interest 5 to 35 C it is exactly the same.
The maximum degree differential range shown (-10 to 50 C) is 0.039 to 0.042.
The average mV/C per temp decade:
-10 : 0.039
0 : 0.040
10: 0.040
20: 0.041
30: 0.041
40: 0.041
And it appears that over this entire range, if you used the average value of 0.040 mV/C the largest deviation would get anywhere would be 0.05 C, and the largest degree-to-degree variation 0.01 C.
From this I conclude that the output is effectively linear in this range, below the error of a typical voltmeter. Of course the voltage needs to be boosted about 100 fold for most milli-range voltmeters.
careysub
Reged: 02/18/11
      #5945975 - 06/29/13 03:57 AM
OTOH - significantly more work seems necessary to incorporate a thermocouple into device that actually reads temperature from multiple probes.
Looking on-line I see lots of sources descrbing how to make a thermocouple thermomenter, that focus on making the thermocouple probe itself, which can be bought for a few dollars, then describes using equipment costing thousands of dollars on a lab bench to actually use it. Rather topsy-turvy in emphasis.
The 1-bit technology at least has ready made components, wiring, and free Linux software designed to monitor tempertures.
The LM-35 IC package puts out enough voltage to monitor directly with a half-decenbt multimeter.
To use the thermocouples effectively it looks like getting into Arduino building and programming (I an an assembly programmer from way back so this hardly fazes me, but it is more work).
BTW - isn't compensating for the wire resistance of the wire from the thermocouple to the amp a problem?
tjmck
Reged: 10/06/07
Loc: Winnipeg, Manitoba, Canada
      #5946087 - 06/29/13 08:03 AM
Look at Adafruit Industries. They have all the solutions you are looking for. If you want to go with K type thermocouples - $10 per, DS18B20 - $4 per, TMP36 - $2 per. Most popular controllers including Arduino and Rasperry Pi, Thermocouple Amplifier MAX31855 breakout board. They also have a data logger sheild for Arduino that incorporates a real time clock and SD card for data storage. Most importantly, they have tutorials on how to setup all of the items and a support community to solve issues.
As for measurement precision, a Class 1 K-Type thermocouple is only rated 1.5C Accuracy from -40C to +375C. For the DS18B20 0.5C Accuracy from -10C to +85C.
There a lots of ways to do what you want. I recently assembled a lab freezer temperature mapping system using an Arduino, Data Logger Shield and 16x DS18B20s that was programmed to take measurements off all 16 sensors every 6 seconds. The data was stored as a CSV file and easily read into a spreadsheet. You could also add a LCD and keyboard or touch screen and even run the whole thing from batteries if you wanted.
Reged: 06/16/04
      #5946429 - 06/29/13 11:42 AM
In general the amp will have a high input impedance so current flow in the wire should not be a problem. I buy my thermocouple wire in rolls and make my own...
How much accuracy are you hoping for?