Hello everyone,
in this period im star testing all my telescopes. In this 3d Id like to submit the output of Winroddier for my Dobson Orion 14xxg, that are in my opinion worth discussing.
Orion claims that the 14xxg is diffraction limited, so I understand it should have a minimum Strehl of 0,8 or lambda/4 P-V. I ran a cycle of 5 tests, in different nights, with different seeing and different stars. In all, I used a ASI174MM with Astronomik green filter. The best result is Strehl 0,85 and lambda/3 P-V (not considering coma), that becomes 0,81 putting in all the aberrations; the worst, shows 0,71 Strehl and lambda/2.3, that becomes only 0,64 and lambda/1,6 adding coma. The results of the other runs are intermediate between these values, implying that the scope barely meets the "diffraction limited" requirements.
Other considerations:
- the scope was always exposed outside for several hours, before testing, to achieve thermal equilibrium;
- astigmatism is very clear also in visual for a trained eye. The axis of the aberration is not parallel to optical axis of the scope, forming with it an angle of at least 30 degrees;
- the main mirror has no pinches which can affect the test;
- i never can achieve a perfect collimation in both sides of focus, i mean if in extrafocal the secondary shadow is centered, it is off center in IF, and vice versa;
- the instrument is obviously suitable for deep sky but in lunary and planetary imaging details become confused, being overcomed in contrast by my C8HD that is much more obstructed;
- I never can eliminate from Roddier output a certain and not negligible amount of coma, despite I ve always tried to be as close as possibile to optical axis.
I attach the 4 Roddier output for your reference (in this order: best without coma - best with; worst without coma - worst with).
What do you think about all this? Could the sample be considered of average quality, or I could open a claim to Orion?
Edited by Vinny1980, 02 April 2018 - 08:13 AM.